Question: Why do raw RGB values differ from calibrated RGB values? I thought calibration just set pixels to wavelength (nm)?

troyb is asking a question about spectral-workbench: Subscribe to answer questions on this topic

troyb asked on June 25, 2018 08:50
110 | 1 answers | shortlink


Here are the first lines of spectra 137547.

Calibrated:

| Average| Red| Green| Blue| | 12.7| 7.65| 10.2| 22.95| | 12.7| 7.65| 10.2| 22.95| | 12.7| 7.65| 10.2| 22.95| | 12.7| 7.65| 10.2| 22.95| | 12.7| 7.65| 10.2| 22.95|

| | | | |--|--|--|

Uncalibrated:

| Average| Red| Green| Blue| | 15| 10| 12| 23| | 15| 10| 12| 23| | 15| 10| 12| 25| | 14| 8| 11| 24| | 14| 8| 11| 24|

The small differences could matter for my daughter's project, measuring a greenness index (green minus blue) to estimate whether algae are growing, especially after we had to recalibrate.



spectral-workbench

question:spectral-workbench



10 Comments

Could not find the actual spectra from the info provided. How about an explicit link or post the graphic?

Is this a question? Click here to post it to the Questions page.


Ah, the table makes the values hard to see. Note for instance that a raw value of 10 scales to 7.65 while a raw values of 23, 24 and 25 scales to 22.95. This happens with all my spectra it seems. That particular spectra is at: https://spectralworkbench.org/spectrums/137547 (and this format can be used to get to any saved spectra).


Also, I didn't save any spectra with saturated channels that I recall, but the recent spectrum https://spectralworkbench.org/spectrums/137643 illustrates that the RGB channels retain the expected integer value only when saturated (value = 255).


Ok, visualizing the spectra helps. 1 - The calibration (wavelength calibration using a CFL source) does not require an amplitude change (though SWB is likely doing this). 2 - Using a Calibration result doesn't alter a measurement spectra's amplitude -- it only corrects the relationship between camera pixels and wavelength (nm). 3 - Once calibrated (wavelength), additional measurements can (should) use that Cal reference. 4 - Remember, the wavelength calibration is only valid if the mechanics of the spectrometer doesn't change (this is a mechanical issue where it is best to leave the spectrometer physical device mounted so it doesn't change for all measurements which are based on that calibration.


Also, 5 - Use 'absolute' measurement values when mathematically comparing amplitude of spectra -- where the only change between plots is the experimental condition being measured.


So, to clarify: what I'm saying is that there seem to mathematical artefacts in the calibrated intensity values. This means that at least for exported csv files, point 2 above is not quite right. The changes to raw intensity are small but I can't figure them out (tried to fish around for the code, but couldn't find that part). So I think this leads to the conclusion there is small bug with unknown consequences (output of code differs from intent), and it will be best remove the calibrations and use raw values when exporting and processing data for quantitative purposes?

Is this a question? Click here to post it to the Questions page.


I'll admit I'm not following your comparisons -- but a few thoughts anyway in case they will help. The second plot at least shows a CFL spectra; although 1) some 'tuning' of the hardware might gain more resolution and 2) the RGB channels are clipping. Clipping adds significant error to the RGB-combined 'avg' result which will likely swamp most other errors. I'd suggest the following sequence: 1) mechanically stable setup, 2) CFL capture w/o clipping, 3) wavelength-cal using the CFL plot. Then... take reference spectra using broad-band light (i.e. through water for example) and then, take spectra with test sample. The wavelength cal's output data (which is applied to other plots as the cal reference) only uses the pixel/nm relationship, not any amplitude data from the CFL cal plot. The difference between the 'water' reference spectra (CFL-Cal'd) and the 'test sample' spectra (also CFL-Cal'd with the same CFL nm/pix data) will be your amplitude difference result and will not contain any amplitude factors of the CFL plot.


On using only raw data (eg. using the CSV 'avg' data): Yes, you could implement a simple linear mapping of pixel:nm with the data from a CFL spectra. Then (and assuming the hardware remains unchanged) apply that mapping to all measurement spectra. Finally, assuming no clipping of RGB channel data, you could perform the difference (water reference vs test sample) -- all using the 'raw' data. The SWB 'calibrate' feature is a convenience; it's not mandatory to doing the spectra processing in another space like a spreadsheet.


Our spectrometer is pretty amazing. Your comments are right on, I certainly recommend avoiding clipping (the warning in SpectralWorkbench works really well). The spectrometer's calibration was very sensitive to bumping – and stoft's explorations of harder builds look really useful. It was really useful to do the calibration, and then just take the wavelengths from the saved and exported calibration spectrum.


Hi, @troyb - we do have a new Lego enclosure which is a step closer to the kind of rigidity @stoft has demonstrated in wood -- we're actually phasing out the older enclosure in favor of this very robust Lego version, and we'd be happy to offer you an upgrade kit at a discount if you're interested?

Is this a question? Click here to post it to the Questions page.


Log in to comment

1 Answers

Thanks very much for the prompt and numerous replies. I think they add up to an answer. There is a small difference in the intensity of the calibrated values, relative to raw uncalibrated values. I made the attached figure to show this. I gather from the replies from stoft that this isn't intended. The difference is small, but I think it may be largest at low values (less than 10 on the scale of 0 to 255).

I also included our experimental results in the attached figure as a second panel.

I'm pleased to say that I've now calculated results and the experiment worked very well. We measured algae with and without an alum treatment that removes phosphorous from the water. The alum makes a flocculent in the water, and our measurements (at 90° to a 20W cool white LED light source) distinguished the 'greenness' of algae (chlorophyll reflects green light, absorbs blue) from the higher reflectance of alum. I suggested we use a normalised difference index (green – blue)/(green + blue) after subtracting baseline (violet).